Discussion Category: Ion Chromatography
calibaration for anion and cation std
what is the parameter for writing program for calibration std in ic 5000
Discussion Category: Ion Chromatography
what is the parameter for writing program for calibration std in ic 5000
Asked by
Dear Sir,
what type of calibration. linearity calibration of anion and cation
yes sir
Dear Sir,
Thank for you reply,
Prepare the standard for 5,10,20,30 ppm for anion or cation.
Create the program, method and Sequence as per the column manua;ls.
Inject th the standard from low to high.
Can i know which software version you using now, whether chromeleon 6.80 or 7.1
chromeleon 7.1
very easy to calibrate sir
Create the program, method and Sequence as per the column manua;ls
can you tell me the steps if i use AS19 4MM
Eluent :20 mM of KOH
Flow rate :1 ml/min
inject volume :25ul
Temp:30C
Suppressor currenr 100mA
Run time :20 min
Page 1 of 1