Discussion Category:  Ion Chromatography

25 ul loop size

If I'm loading a 100ul sample volume on a 25ul loop, does that mean that the sample volume goes thorugh 4 x

avatar placemark

Asked by

gordy
Report this Post
cloudyarunReply by
cloudyarun

No. After filling the 25ul loop rest of things goes to waste

 

Be happy  
Report this Post
avatar placemarkReply by
gordy

Thank you for your quick response.

However, when I load a 200 ul sample volume my peak area is 2x the size of that of 100ul.

If the rest goes to waste I should not be seeing my peak area double?

Report this Post
cloudyarunReply by
cloudyarun

Do you have autosampler or manual injector/.

 

Be happy  
Report this Post
avatar placemarkReply by
gordy

autosampler

Report this Post
cloudyarunReply by
cloudyarun

i need autosampler Model . and see the loop size in autosampler and injection valve in DC

Be happy  
Report this Post
avatar placemarkReply by
gordy

The Autosampler Model is Dionex AS, loop sizes are 25ul

Report this Post
cloudyarunReply by
cloudyarun

Definitely there should not be such kind of problem. I think you should check the loop once again or check the injector linearity.   Also, did you get reproducibitly for the standard?

Be happy  
Report this Post
cloudyarunReply by
cloudyarun

that means autosampler not taking the exact injection volume

Be happy  
Report this Post
avatar placemarkReply by
gordy

Thank you ...I'll take a better look at the autosampler/settings

Report this Post
avatar placemarkReply by
Bob Kafato

I think you have to program into the auto sampler the size of your sample loop if you have this kind of problem.

Report this Post