Discussion Category:  Ion Chromatography

Fluctuating Pressure

 

The pressure on our unit is fluctuating between 1200-1700 and is constantly jumping 20-30 psi.  Every time it makes a knocking sound it drops by about 300 psi for less than a second. Any suggestions?

 

 

avatar placemark

Asked by

Candice B.
Report this Post
avatar placemarkReply by
cemchem

I had something like after instrument had many months of use.  This might happen with bad check valves, bad pistons or piston seals, but mine was a bad bearing in the pump drive module.  It is expensive to replace the pump module.