Discussion Category:  UV/VIS Spectrophotometers

Wavelength Correction

Wavelength Correction is not executed normaly. in the measurements, the spectra is shifted to the right.

avatar placemark

Asked by

AnF
Report this Post
avatar placemarkReply by
JRB

There are several causes for a wavelength shift, but basically the sine bar driving the monochromator needs to be adjusted. There is most likely an underlying cause why the shift has occured, which would need to be investigated to ensure that the problem is resolved. This would mean cleaning and lubricrating threads in the sine bar and the monochromator carrier, followed by recalibrating the zero order and D2 line offsets. In hte first instance you should contact your local JASCO representative who can help you with this.

Report this Post
avatar placemarkReply by
Lucille F. Parham

I had found that the information is very helpful.

Report this Post

Page 1 of 1