I would like to know if users are happy the the H-Class for UPLC analysis
avatar placemark

Asked by

KHunter
Report this Post
LabWrenchManagerReply by
LabWrenchManager
We have a couple of systems - standard - not H-class. Depends on what you want to use it for - if you want 4 bottles on top - quartenairy solvent system etc - you will need the H class, otherwise go for the standard one. But we have no problems with our systems and staff are happy with them. This response was provided to LabWrench by Nico van Loon

Melanie Playne
LabWrench Product Manager

Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
 
Report this Post
LabWrenchManagerReply by
LabWrenchManager
The two main differences, that I see, between the Classic and the H Class are Binary vs Quat and residence time. If you don't need Quat capability and need extremly fast methods go with the Classic. H Class has more flexibility at the expense of an additional 1-2 minutes per method. For both instruments, make sure you optimise your band width (called resolution for some reason) and your sampling rate. A sampling rate of 5 gives 2-3 times the sensitivity as a sampling rate of 20. This response was provided to LabWrench by Theresa Wilson