Beckman rotor moves with difficulty
The rotor movement is difficult when try to swirl manually. When started the run, the centrifuge shows error message and it stops.
Discussion Category: Centrifuge
The rotor movement is difficult when try to swirl manually. When started the run, the centrifuge shows error message and it stops.
Hi
Arshad, You misunderstand me. I was not worried about you removing the rotor from the drive assembly, I am concerned about the fact that you said that it is difficult to SPIN the rotor by hand and that should not be the case with that drive motor. There is no way to really lubricate the bearings in that motor as everything is sealed up so that is why I believe that you are looking at a drive motor replacement.
Don
Thanks Don
It would be expensive to replace the motor. Is there somewhere it is cheaper than getting from Beckman?
Arshad
I am located in Montreal, Qc, Canada.