Discussion Category: Scintillation Counters
Zero counts
Trying to resurrect an LS 6500. Got the elevator working and fixed a printer problem. Now I'm getting H# Count rate too low error when counting. If I switch to just plain cpm I get a count rate of zero. Suspect power supply, but everything else seems to work. Is there a separate HV supply for the PMTs? Any way to test that? I've got lots of instrument troubleshooting/repair experience, just no schematics or service manual make it tough to proceed.
SN is 7067355. It's got the Wyse thin client.
Thanks for the info. System is surprisingly clean and dust free.
There are two LEDs on the HV supply. DS1 lights when counting and DS2 flashes briefly then stays out. If I disconnect the cables to the PMTs both LEDs stay on. If I swap the cables DS2 lights, and DS1 does not. So the problem seems to be on the left side: either cable, socket, or PMT. I'm tempted to dig into the detectors and see if I can swap things there amnd/or hopefully see some obvious problem, even if there's not much I can do about it.
How can I tell if I've got the Singles option? Would that be listed in the System Info page?
The HV supply is working fine now, with both LEDs on steady, and a voltage of about 1100 (measured with HV probe). I've been watching the LEDs on the coincidence/analog board; DS4 lights when it counts, but DS2 doesn't. After swapping things back and forth between Left and Right it sure looks like the pre amp/base is the problem. I've got a line on a replacement, but will try cleaning the old one first. Apparently the left side gets dirty from the fan on that side and can cause problems.
I'm not seeing a singles option anywhere, unfortunately.
I never was able to "fix" a preamp by cleaning it. Replacement is generally what you will be looking at but definitelty give it a try and you are correct in that just about everything on the left side gets messed up because of the air flow although that fan runs so slow that it really did very little cooling and I found more than a few poorly maintained LS counters with non functional fans still functioning properly. One thing you may find that I dealt with on a regular basis is that the closed cell foam around the PMT's has degraded to the point where it really no longer acts to support the tube within the shielding to help keep the tube centered. If you do find that it has been squeezed flat then I recommend that you replace it as this can have an impact upon the overall counting efficiency.
Don
Thanks for the tip about the foam. It's looking a little tired, so if I do get things working I'll be sure to replace it.
If I can think of anything else then I will let you know. It is has been around 5 years since I actually worked on one of these so talking about it is dredging up old memories.
Don
Your advice and help is greatly appreciated. Service engineers who are willing to share information with customers are worth their weight in gold; not all companies allow/encourage it and not all engineers are willing to do it.
Rob
Are those the connections on the socket side of the board? if so, it looks like a tight squeeze to get in there and solder them, but I can see how that might help.
Rob