Discussion Category:  Centrifuge

Epopendorf 5415R won't turn on

I have looked at the various answers over the issue of the centrifuge not turning on after it has been turned off. The 5415R control board is different than the 5415D. I have tried to replace the the 2.2uf capacitor to a 10uf capacitor as explained without any success. Also the four 68K resistors are replaced with 8 ohms and add up to 32 ohms. The same situation happens when the ribbon cable is disconnected while the power is on, it will reset. Does anyone have any ideas!

Thanks for your help

Guri

 

avatar placemark

Asked by

Guri
Report this Post
SaharaSimbaReply by
SaharaSimba
Hi Guri... I can assist with repairs on the board if you send it to me. And will work with you if you are on a tight budget. See my info www.saharasimba.com
Report this Post
avatar placemarkReply by
Tech
Did you ever fix your issue? The capacitor needs to be at least 22 uF 35V.
Report this Post
avatar placemarkReply by
NormW

Hello, yes Tech is right. It has to be a 22UF 35V.  This has worked for me a couple of times.