Discussion Category:  ShakersRefrigerated Incubator

Can't open front door (latch broken)

Can't open the front door of unit. Latch is broken. Anyone know how to open the front door of this device?

avatar placemark

Asked by

BGantt
Report this Post

1 Reply

avatar placemarkReply by
WarwickJon

Quote BGantt

Can't open the front door of unit. Latch is broken. Anyone know how to open the front door of this device?

Hi, 

Bit of a fiddle this one.  The only way to get the door open is to get at the nut on the backside of the locking bolt. 

This is situated at the bottom of the chassis, in a central position along the bottom.  You will have to remove the bottom grill and then push your way through a little insulation possibly, with a spanner to remove the nut.  Once you have done this you should be able to open the door.

The latching mechanism itself sits within the door casing.  You will have to remove the door to get access to this.  You can buy replacement parts from Eppendorf (who now own Innova).

I hope that helps.

Jon

Report this Post