Discussion Category:  GC Systems

Front Inlet Septum Purge Shutdown

After initiating the method, the GC septum purge flow refuse to attain its set point of 3ml/min. Septum Purge flow remain stagnant on 0.1ml/min without increasing any further, and eventually shutdown after a few minutes.

Please what could be wrong?

Please find below the set points for my GC with the actual (current) readings indicated in parenthesis.

Front Inlet Parameter

Mode= split;

Temp = 150 degC;

Pressure= 23.97 psi;

Split ratio= 10;

Split flow= 34.69ml/min (actual= 37.56)

Septum purge = 3ml/min (actual= 0.1ml/min)

Gas saver ON

Saver flow =20 ml/min

Saver time = 2 mins

Note that when i disable the septum purge, the GC does not shutdown but as soon as i put the septum purge ON, the GC shutdown in safe mode.

avatar placemark

Asked by

ADE
Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie
It is likely the EPC module. As I recall the only calibration option is to zero out the flow for the septum purge, if that doesn't work I would look at replacing it. Dan

Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
ADE

Thanks Dan!

The septum purge is working fine now after the EPC was changed out