Discussion Category:  Titrators

RS232 Data export

Hello,

I have been trying to set up this selecting inside peripherals:

RS232 Data export

- MB/COM1 or also  MB/COM2

- Speed: 9600

- Data bits: 8

- Stop bits: 1

- Parity: None

- Handshake: None

However, I don't have any print option button from results or methods...

Does anyone know if this could work?, maybe a license is required.

Thanks a lot.

 

 

 

 

 

avatar placemark

Asked by

Leo_tech
Report this Post

Replies

jburkman11Reply by
jburkman11

To activate the export need to select the folowing setting in the instrument's setup and only export when the method is performed:

Printer type- 'RS-232 data export'

The Summary parameter in the 'Record' method function

'Record' inside the loop: 'Per sample' or 'Per series'

'Record' outside the loop: 'Yes' or 'No'

Note: The other parameters which are selected in the 'Record' funcion are ignored.


Jennifer Burkman
METTLER TOLEDO
www.mt.com