LN2 level sensor for quantachrome surface analyzer

Dear colleagues
I have a problem with quantachrome surface analyzer type 3200E.
The liquid nitrogen level sensor (LN2-RTD) does not detect the rise of the dewar. therfore the analysis is aborted.
Is there any solutions to control the rise of LN2 dewar in manual mode.
Regards

avatar placemark

Asked by

Fouad-DZ
Report this Post

5 Replies

avatar placemarkReply by
DonMerrill

Contact your local Anton-Paar Service representative for a replacement probe.  The RTD probes are no longer manufactured.  You'll have to install a Hybrid-CLS probe instead.

Report this Post
avatar placemarkReply by
Fouad-DZ
Thank you for your reply;  Could you please send me the details of Hybrid-CLS probe????   Regards
Report this Post
avatar placemarkReply by
DonMerrill

AP 195455     Level Sensor, Hybrid Type, 77 K, Long

Report this Post
avatar placemarkReply by
sales.chem7561

Dear DonMerrill  who is the manufacturer of AP 195455     Level Sensor, Hybrid Type, 77 K, Long?

we need to buy one sensor and I can not find it online. 

Report this Post
APUKReply by
APUK

www.anton-paar.com 

 

If you go through the website the online assitant will help you locate your local contact. 