Discussion Category:  Centrifuge

EPPENDORF 5804R ERROR 1 ALL THE TIME

EPPENDORF 5804R Centrifuge, every time we want to spin something down ERROR 1. 

 

PLEASE HELP

avatar placemark

Asked by

BaeLab103
Report this Post

6 Replies

avatar placemarkReply by
WarwickJon

Error 1 reads an error in speed measuring system.  The manual suggests trying a different rotor.

 

https://www.nist.gov/system/files/documents/ncnr/Eppendorf_5810R_Centrifuge_Manual.pdf

Report this Post
avatar placemarkReply by
BaeLab103

Yes, it doesn't happen with another rotor. But I need to use this rotor, so what do I do?

Report this Post
avatar placemarkReply by
WarwickJon

Is the rotor you are trying to use approved for use in this centrifuge?  If so, the unit isnt recognising it.

Report this Post
avatar placemarkReply by
BaeLab103

Yes, it is approved for use in this centrifuge. 

What can I do to make it recognize it / properly measure the speed?

Report this Post
avatar placemarkReply by
WarwickJon

If the rotor is a newer type maybe the firmware needs an update?  May be worth checking with Eppendorf.

Report this Post
avatar placemarkReply by
BaeLab103

Thanks, will do.