Discussion Category:  Hot Plate Stirrers

Error E 9

I have used cmag hs 7, and need solution for error 9. May be in this forum any some solution.Thanks
avatar placemark

Asked by

taufikmas
Report this Post
avatar placemarkReply by
septianitia1387

How to repair hotplate IKA C-MAG HS 7 with error code E9. Thank you very much>

Report this Post
avatar placemarkReply by
mohammad.habibi824873

Hello,

Is there anyone know the answer of above question?

please share the answer.

Report this Post
avatar placemarkReply by
s.zangi

hello

Were you able to fix the device with this error?

Report this Post
avatar placemarkReply by
s.zangi

hi

I also have a device that has the same error. Could anyone fix it?

Report this Post
avatar placemarkReply by
nbladesclarkprorx
03.03.2023: I had this Error code today. I found that if the protective data accessory "plug/.cover", or a data cable, is not attached then I get the E. 9 (error code 9). I tested this theory by removing the data plug, I got the error. When I put it back in and turned the hot plate off then back on, the error code was cleared.
Report this Post
avatar placemarkReply by
miguelgomez5963
Have had this error for quite a while, are there any suggestions on what to do? Thank you
Report this Post
avatar placemarkReply by
nbladesclarkprorx
miguelgomez5963: please see my reply in this thread posted on 03.03.2023 (nblades)
Report this Post
avatar placemarkReply by
SDS
nbladesclarkprorx: I got the same error but I don't have a data plug. Could you please advice what is the electrical schematic of this plug? Can I put jumpers on the defined pins of this connector instead of the data plug?
Report this Post
mgoneeReply by
mgonee
hello. Error code E9 is not a fault condition indication. It can be resolved by connecting a temperature sensor to the 6-pin connector on the back. If you don't have a temperature sensor, connect the two terminals with a 3-5 ohm resistor. The two terminals are at 12 and 4 o'clock when the tip is at 6 o'clock. Sorry for the lack of explanation due to my poor English. ^^ Good luck.
Report this Post
avatar placemarkReply by
SDS
mgonee: I really appreciate your help! I followed your advice connected the terminals for the temperature sensor located at the back of the device as you suggested and the E9 error is now resolved. Thank you!
Report this Post