change lamp but message warning low light

hello

I changed the lamp on our akta explorer but I still have the warning low light message.
I want to go to the mode service but I can't find where to put the code.
to reset the lamp hours
thank you for your help

best regards

 

andré

avatar placemark

Asked by

andrel
Report this Post
avatar placemarkReply by
miguelR

Hi, 

 

still problems with that ? if so:

break the connection to pc /r estart UV detector and turn detector knob anti clockwise until service code is required./ Then you have to insert service code (I think it is 1996 (not sure. I have to double check that). Use the knob to enter the code). 

when in the service menu, check for the options.  Intensity tests , I think there is also a reset you can do. 

But you might as well have a burned opfical fiber or dirty flow cell ? 

 

Bye and good luck, 

Miguel 