Hi . Does anyone know what Fatal Error is? : 0x8024 Data: 0x00,0x00,0x00,0x00 . It would be good to know if someone has compiled a list of these errors and there meaning?

 

avatar placemark

Asked by

Labcare
Report this Post
avatar placemarkReply by
Bodasafa

0x8024 means the lid has overheated exceeded 115C. Run the heated lid diagnostic.

Report this Post
avatar placemarkReply by
mohammadtabrizie1354

HI all

dear experts(specially dear sir BODASAFA):

WE WERE SUPPOSED TO UPDATE THE FRAME WARE OF OUR VERITI PCR BUT WHEN WE DO IT AND RESTARTING THE DEVICE IT CAN NOT BOOT UP COMPLETELY AND STAYS AT THE ""FINALIZING INSTRUMENT BOOTUP" SCREEN

WE NEED YOUR HELP

THANK IN ADVANCE