New Brunswick C25 Temp. Issues

We have a New Brunswick C25 Incubator Shaker. The temperature on the LED will continually read 38.7 degrees Celsius even when the temperature reaches over 60 degrees. I assume it is a bad temp probe but wasn't sure if this was a common issue and if someone could confirm.

avatar placemark

Asked by

jdelaney@alt-inc.com
Report this Post

1 Reply

avatar placemarkReply by
Ed S

I've had a couple of these units that have been modified by revious technicians who replaced the temp set thermostat with an OMRON temp controller and an electronic relay. The controllers I found to be stuffed so I replaced them and presto all working.