Discussion Category:  Centrifuge

error code

will not start

avatar placemark

Asked by

Trekky
Report this Post

Replies

avatar placemarkReply by
dpkleessr
Did any error lights show up on the display panel? Saying it won't start does not tell me much to help you. Don
Report this Post
avatar placemarkReply by
Trekky

Good Morning Don:

To clarify i should have said the J2-21M PROBLEM began when the operator informed me that the door would not close. I got the door opener to work. Next day i was suprised when the Speed error poped up. I did reset the swich on the board and 220 came up and hit it again and zero was indicated. Same problem. I even swapped the entire set of the four pcb board from another machine and still no Go. Any suggestions??

                                                    Bob J.

Report this Post
avatar placemarkReply by
Trekky

Don:

My problem is now resolved. I found a 0.75 amp fuse F101 was blown.

I had gone to the trouble of exchanging the entire chassis and that is how i found the problem.

I am testing the J2-21M and all is good.

Thanks for your assistance

Bob J