Discussion Category:  DNA Synthesizers

Software or Service Manual

Anybody know if this unit has some software that might work or if there is service manual? The unit we have has two lines on the display but nothing else. I cant diagnose if the issue is with the board or the display board. Any help is appreciate it.
avatar placemark

Asked by

MedRun
Report this Post
avatar placemarkReply by
lyndamillstine

Hi, Your problem is most likely due to the Controller board or the display itself.  There is a knowlegeable company that services this instrument.  Contact ap@anteris-systems.com  Anteris systems should be able to help you.

Report this Post
avatar placemarkReply by
ivyp1
Hello, we have a similar issue. Were you able to find a resolution or a service manual?
Report this Post
avatar placemarkReply by
MedRun
The issue is the software card. It has a battery that died and sadly it need to be replaced. The entire card not only the battery, once the battery is gone the program is also gone. Search google there is company that sales them in California, for about 180$
Report this Post
avatar placemarkReply by
labotronics.mary
Start by checking the power supply voltages and the cable/connectors between the main control board and display board, since two lines on the display can indicate either a display initialization problem or a controller fault. If the power and connections are correct, the service manual or wiring diagram is the safest way to test the display and board separately. Contact the manufacturer or an authorized service provider with the exact model and serial number to obtain the manual and compatible software. Avoid replacing the boards until the supply and communication signals have been verified.
Report this Post

Page 1 of 1