Unable to set carrier flow at 500 ml/min

Hello, After having to shut down the lab for the last 7 months (no students) I'm slowly starting the lab up. One of the task was to ensure the lab equipment is working. I have a Shimadzu SSM-5000A (solid module) attached to a TOC-L analyzer. After the initial and routine checks, I tried to start the system up. One of the very first steps is to open the high purity O2 tank and set the carrier gas supply pressure to 44psi . The second step is to set the carrier gas pressure to 200kpa (29ps) using the regulator valve in the SSM-5000a front pannel.  The third step is to set the flow rate for the SSM-5000a to 500mL/min. This is where the problem occurs. It doesn't matter which way I turn the mass flow controller knob the flow gauge does not change from ZERO. I have verified that the gas supply is adequate but I'm at a loss on what is going on. Can anyone help?
avatar placemark

Asked by

altsolrac1672
Report this Post
avatar placemarkReply by
altsolrac1672

After following the the flow path from tank to mass flow regulator I found out the problelm as a stuck floater ball in the mass flow gauge.  After a few taps it worked normally.  