"L7 Door problem during run...."

Hi, My centrifuge is Avanti J25I, the message "L7 Door problem during run call the service " is coming up. Ant key look likes locked. Please tell what happened? I appreciate! Kate
avatar placemark

Asked by

kate8716
Report this Post

1 Reply

avatar placemarkReply by
dpkleessr

Not sure if you got this repaired yet but if not then your centrifuge is telling you that one of the door latches is not closing properly or sensing that it is indeed closed.  It could be the latch hanging up or the sensors.  Sometimes cleaning and relubricating the latches fixes the problem but difficult to tell without actually seeing the unit and the latches.

Don