Discussion Category:  PCR / Thermal Cyclers

Bio-Rad Laboratories, Inc. - CFX96 Block, C1000 Base, Post Error x000B0035,2:

Bio-Rad Laboratories, Inc. - CFX96 Block, C1000 Base, Post Error x000B0035,2: The unit passes the diagnostic tests in the Admin settings. Has any any seen this or fixed this before?
avatar placemark

Asked by

Jkirk
Report this Post
avatar placemarkReply by
drogo_khal

shuttle needs to be replaced whenever this happens.

Report this Post
avatar placemarkReply by
Jkirk

Drogo Khal,

Is the shuttle is in the CFX96 Block?   Are there any parts that can be ordered?  Who is a good repair vendor, if I cannot get parts?

Report this Post
avatar placemarkReply by
drogo_khal

yes, the shuttle is the item that incidents light in each well of the pcr plate. It is in the block (the upper part of the machine). It can usually be bought,we buy it. I am not in the US, so that know any repair vendor. You should reach out to your supplier.

Report this Post
avatar placemarkReply by
Jkirk

Drogo Khal,

 

Could you send me the part number that you use.  I will see if I can find it here.

Thanks,

Jkirk 

Report this Post
avatar placemarkReply by
drogo_khal

You may try: 10013071. Not sure if this is still useful, havent been here in a while.

Report this Post