Hi, on a V30 titrator (KF), the solvent manager will not fill or drain anything. All connections have been checked, no leaks. What can I do to fix that issue? Thanks, CJB
avatar placemark

Asked by

Bassticular
Report this Post
jburkman11Reply by
jburkman11

When you turn the solvent manager pump to drain or fill, the the reagent move any amount in the tubing? If yes, even moving a few centermeters, then the pump is working and there is an air leak somewhere. The solvent manager pump is a very low pressure volume pump. The littlest of fitting air leak will not allow the reagent to drain or fill. Also, do you feel and vacuum at the small port on the top of the solvent manager pump when draining? Or can you feel any air flow from the bottom port on the solvent pump during a fill operation?  If there is no movement of the reagent in the drain or fill lines, then the solvent manager pump will need to be replaced.


Jennifer Burkman
METTLER TOLEDO
www.mt.com  
Report this Post
avatar placemarkReply by
islampes93449
Same problem man Any solution ?
Report this Post
avatar placemarkReply by
Bassticular
Hello islampes93449, it turns out someone overfilled the waste bottle and the waste solvent got into the pump housing. I had to take apart the pump housing and dry all the internal components. Make sure you don't have any air leaks at any of the connections.
Report this Post
jburkman11Reply by
jburkman11
Unfortunately, most likely even after cleaning of the solvent manager pump that will need to replace.

Jennifer Burkman
METTLER TOLEDO
www.mt.com  
Report this Post
avatar placemarkReply by
Bassticular
@jburkman11 Actually nope, after cleaning out the solvent from inside the pump housing & reassembly it works great!
Report this Post