Increase safety when using Hydrogen
I would like to use Hydrogen as the carrier gas in my HP 5890, but I am afraid of the risk of explosion, are there any modifications I can make to the 5890 to increase safety when using Hydrogen?
WrenchGuy
Neither of my 5890s have sensors. I've used hydrogen carrier on all
of my capillary GCs for over 12 years without an incident. About a
decade ago, Walt Jennings visited me, and as I was showing him
our 0.1mm columns used for triglycerides and sterols, and the mass
flow controllers we put upstream of our older GCs that just had pressure
control. He said that the only hydrogen explosion that he had ever had,
occurred when he forgot to block off a 1/4" packed column injection port
fed just by a pressure regulator, and hydrogen was flooding into the GC :-).
Virtually all modern GCs have sprung loaded doors, so that if an explosion
does occur, the over doesn't disintegrate with debris hurtling around the
laboratory. I would recommend using hydrogen all the time, unless you
have a very good reason for not using it, it's faster, provides better
resolution ( only nitrogen at 10 cm/sec beats it ), and is cheaper than
helium.
We use industrial grade hydrogen as carrier, passed through
indicating drierite/molecular sieve traps, followed by an oxy-trap,
followed by a indicating oxy-trap, followed by an activated carbon
trap. Our three capillary GCs ( two with liquid autosamplers ) are
in use nearly 24 hours/day seven days a week - only closing
for Christmas/New Year break, at which time every year or so,
we regenerate the driers and oxy-trap. Most of our columns last 4-10
years, the trick is never to allow a column or injector to be exposed to
air at above-ambient temperatures, and to use a packed split liner to
collect all the junk. We try for minimal sample preparation ( grinding
up meat in chloroform and methanol, 0.45um filtering and injecting
the solution for triglyceride and sterol analysis ), and the columns
will last for years. Using industrial grade gases is also a lot cheaper
and they tend to be fairly pure ( for detectors, the gases pass through
drierite/molecular sieve and activated carbon traps only ). It's
worthwhile having a purge valve on the high-pressure cylinder manifold
to purge any air that gets into the hose when changing cylinders - and
we keep all lines from cylinder to GC toggle valves pressurized at all
times ( they all have pressure relief valves at around 400psi should
the regulator leak through - bursting disks can also be used.
The first thing I do to new GC is to mount Swagelok on/off toggle valves
on the carrier gas lines as close as possible to the controllers ( we
plumb both Helium and Hydrogen to GCs, as some standard methods
still specify He, with a 1psi non-return valve on the He to ensure H2
never gets into the He line if somebody inadvertently opened both
toggles at once ). I ensure all lines are 1/8" copper or SS, with 2-stage
regulators with SS diaphragms, and no plastic anywhere. Using toggle
valves means that changing septa etc. can be easily performed without
affecting settings.
I have no worries at all about hydrogen leaks, we have a Matheson
( actually from Gow-Mac ) leak detector that we use to check any
new plumbing, and when people are learning how to change columns.
I tend to operate at much higher linear velocities with hydrogen so
we can perform rapid analyses, e.g. one GC analyzing free fatty acids
currently has a 10 meter column where the solvent peak is coming
out in 5 secs, and the total analysis time is 2.5 minutes.
So, in summary, provided the GC is designed to cope with a hydrogen
explosion ( spring-loaded door or vent ), hydrogen is an excellent
carrier gas for both manual and automated unattended operation.
There is no need for additional safety devices, but if you wish you
can get sensors, and can install mass flow controllers upstream of
the GC to limit the maximum flow to about 250 ml/min if using
pressure-controlled carrier gases that doesn’t have a flow controller.
SOURCE: http://yarchive.net/chem/gas_chromatography.html
WrenchGuy
Also, you could ensure all split vent ( and septum purge if paranoid ) lines
are metal and lead to a location where they can be safety
vented. I use a long 1/4" copper line from each instrument
to outside the instrument room. Don't manifold the lines
from different instruments.
You can put a metal sinter on the outlet if the vent line
does not go to a restricted ( no flames ) area. But if you
use a sinter, ensure that it is cleaned regularly to prevent
backpressure. I allow the septum purge lines ( 2.5 ml/min )
to vent into the laboratory, but if you have sensors, or a
small room with poor circulation you should also vent them,
and they can be connected to the 1/4" split vent line.
I cut the split vent line on 5890s and put a 1/8" tee in it.
I then put a toggle valve on the line going outside and a
cap on the split vent outlet on the front of the instrument.
This means that when I want to set split flows, I just uncap
the front, add the soap bubble meter, and close the toggle.
I suppose richer people could just buy a three-way valve to
ensure people can't get the sequence wrong and completely
block the split flow, over pressuring the injector if mass
flow control is used.
SOURCE: http://yarchive.net/chem/gas_chromatography.html