Discussion Category:  Centrifuge

L7 and L10

j-26 is throwing codes for L7 and L10. Have gone in and serviced the 2 latches and confirmed functions as well as the door lining up. What else could cause these

Equipment: Beckman Coulter - Avanti J-26S

avatar placemark

Asked by

Tech23
Report this Post
avatar placemarkReply by
dpkleessr
Well, an L7 diagnostic says that the door is unlocked during the run and the L10 diagnostic says that the door interlock status and control status are "inconsistent" during the run which means that they are not matching up for some reason. You say that you "serviced" the latches. What does that mean you did? We have to start there before I can try to help you any further. Personally, it kind of sounds as if the door latch sensor at the turnbuckle area coming up from the foot pedal may have had the cable jump out of pulley which can definitely cause that problem.
Report this Post

Page 1 of 1