Discussion Category: Liquid Handling
Dionex ASE 200
The needle mechanism will not move from the home position.
Reply byLabWrenchManager
I have been working with Dionex ICS2000. Never had experience with this particular intrument. You best choice would be calling Dionex (actually it is Thermo Scientific now if I remember correctly) directly, they purchased it last year October). The company is pretty good with customer service.
Thanks to J. Duan for providing this response to LabWrench.
Melanie Playne
LabWrench Product Manager
Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
Reply byLabWrenchManager
you could check if your gas supply for the system is OK and if the overall system pressure is set to 50 psi, opening the front door of the instrument you may see it showing at a regulator. Also it can be something related with internal solenoid valves or any kind of blockage at the arm movement. Could you perform a rinse? If you need any more assistant you could call Dionex Support Center.
Thanks to M. Chalom for providing this response.
Melanie Playne
LabWrench Product Manager
Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
Reply byLabWrenchManager
I used to work for Dionex and if you call the Salt Lake number below they can offer support on the ASE 200,
1182 W. 2400 S, Suite 8A
Salt Lake City, UT
84119
United States
Phone: +1 801 972 9292
Fax: +1 801 972 9291
Thanks to A. Bellinger for providing this response.
Melanie Playne
LabWrench Product Manager
Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
Reply byLabWrenchManager
you need almost 100 PSI of gas supply for the needle to work. does it actually try to move or just set still? I have to agree with Ann as will Salt lake people know the best about these instrument, we actually have ASE300, that experienced similar problem, and it was the gas supply pressure.
Thanks to M Abouzari DPM for providing this response
Melanie Playne
LabWrench Product Manager
Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
Reply byLabWrenchManager
I too use to work for Dionex. I agree check the gas pressure, should be set between 50-60 psi as I recall. Other than that the folks at Dionex SLTC should be able to help.
Best of luck.
Thanks to D. Jarzinski for providing this response.
Melanie Playne
LabWrench Product Manager
Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
Add your reply:
Page 1 of 1
