The BIO-RAD CFX96 Real-Time System C1000 Touch Thermal Cycler is displaying " System self test...Failed" with Block CT029301 Errors, Error Number: x000B0035, Info:2, Description: Power on test detected a problem. What could be the problem?

Equipment: BIO-RAD CFX96 Real-Time System C1000 Touch Thermal Cycler

avatar placemark

Asked by

hillary.cropnuts4645
Report this Post
avatar placemarkReply by
ZakLabcare
In admin mode, run the diagnostics test and see what errors it picks up and if it goes into more detail. Service manual states: POWER-ON TEST FAILED: ROM ERROR. IF ERROR PERSISTS, CONTACT BIO-RAD TECHNICAL SUPPORT. Bad Sensor, limit switch, loose parts or bad FFC or main board Test all switches and sensors, tighten and align loose screws and parts. Replace bad board