Discussion Category:  PCR / Thermal Cyclers

error in pre-PCR

I have an ABI 9700 with an error in pre-PCR: the instrument could not reach 94°C in 4 minutes then the run was aborted. Do you know why this is happening and how I can fix it? thank you
avatar placemark

Asked by

khouloud
Report this Post
avatar placemarkReply by
ZakLabcare
This is likely from the heated lid not getting to temperature before starting the run, on these older units they can be a bit slow and if you have the lid lifted up or pressed against a plate it really does struggle, my suggestion would be to perhaps lower the lid set temperature or help it get up to temperature by keeping it closed and not pressed against anything.



Zak Purtell
Senior Engineer
Labcare Service Ltd
www.labcareservice.co.uk
Tel: 01376 332909
Service, calibration, repair and sales of laboratory equipment.
 
Report this Post
avatar placemarkReply by
khouloud
Thank you for your response. The issue doesn't appear to be with the lid, as it consistently heats to 103°C before the run, as expected. However, once the run begins, the sample temperature does not exceed 30°C. When we attempted a calibration verification, the lid temperature reached 85°C, but the block temperature remained at 28-30°C.
Report this Post