Discussion Category:  Shakers

Rusted shaking mechanism

Hi everyone, Our lab has miniton HT CO2 shaking incubator which I have been using for the past year or so. Recently, it started making a loud grinding noise while shaking. There was also some brown residue below the shaking plate. Upon removing the plate, I saw lots of rust and brown gunk all over the bearings and the big rotating weight (black semi circle). The bearing seems to have badly rusted which is causing all the noise. Does anyone have an idea about the parts that may be needed to make repairs and a how to guide for the same. Thanks in advance.
avatar placemark

Asked by

as@h
Report this Post
avatar placemarkReply by
GaryTompkins
Hi there, It sounds as though you are referring to the upper swivel bearing assembly. With luck, this is all you'll need. I would need the serial number to get you part information. I'd also recommend lifting the table and taking a look underneath to see the condition of everything. Also, I'd like you to spin the counterweight assembly by hand with the table disconnected to feel the condition of the lower bearings. You can contact me directly at g.tompkins@infors-ht.com
Report this Post
avatar placemarkReply by
GaryTompkins
Hi there, It sounds as though you are referring to the upper swivel bearing assembly. With luck, this is all you'll need. I would need the serial number to get you part information. I'd also recommend lifting the table and taking a look underneath to see the condition of everything. Also, I'd like you to spin the counterweight assembly by hand with the table disconnected to feel the condition of the lower bearings. You can contact me directly at g.tompkins@infors-ht.com