Discussion Category:  Particle Size Analyzers

Particle Analyzer

We are currently evaluating a Coulter Multisizer 4 for possible purchase. Our current instrument is a PSS/Accusizer 770. The Coulter uses the Coulter Principle (Electric Sensing Zone) for counting particles. Our Accusizer uses light scattering and extinction. When running one of our typical samples, Coulter came up with counts that were 3000 times greater (2-10ยต channel size) using their technique. We are having difficulty understanding why there would be that great a difference. Is the sensitivity of the Coulter technique actually that much greater than light scattering/extinction?
avatar placemark

Asked by

CI
Report this Post
avatar placemarkReply by
lew_b
Hi, I would strongly advise against trying to make "absolute comparisons" of measurements made using two particles analyzers whose principle of measurement are different. This is especially true for these two instruments, as they are both making "indirect" measurements. In other words, they do not "directly" measure particle size, they measure something that is [b]PROPORTIONAL[/b] to particle size and [i][b]convert[/b][/i] it to size. Since they are indirect methods, their "measurements" are "calibrated" using standards of known size, and then incoming measurements from your sample are compared against the calibrated "known" signals. If the sample does not have the identical characteristics as the standard (which is almost always the case, since the standards are typically spheres of glass, polystyrene, or other material), then the measurements produced have to be taken with a grain of salt, because the sample's physical properties (light scatter, electrozone signal, etc.) differ from those of the calibration objects. So, it is usually a good idea to look more closely at "trends" as opposed to "absolutes" when using these instruments. The other thing you may want to consider is to "calibrate" your system using your own samples compared to microscope measurements. This way, you would have "known, direct measurements" of your own sample material to compare as a baseline. Since you also state you are interested in 2-10um sizes range, you may also want to consider a "direct" measurement technique such as imaging particle analysis (full disclosure: my company makes such a device!), which is essentially fully automated microscopic measurements of each particle. This has the added benefit in that it give you particle shape information as well (along with an actual picture of the particle!). However, if you are interested in particles smaller than 2um, then imaging will not work due to diffraction limits and sampling theory. Hope that helps! Lew Brown Fluid Imaging Technologies, Inc. (207) 846-6100 [url=http://www.fluidimaging.com]www.fluidimaging.com[/url]
Report this Post
avatar placemarkReply by
patrickspos
Dear CI, My name is Patrick O'Hagan and I work for Particle Sizing Systems, manufacturer of the Accusizer 770 (current model is the 780). While it is true that the Multisizer and the Accusizer use different technologies, they are both particle counters that size and count particles over a comparable range of sizes. They should be expected to produce particle counts that are similar. A difference by a factor of 3,000 suggests something is wrong. As you state that you have a 770, then it must be a tool that is over a dozen years old. Have you verified the performance of your Accusizer system recently? You can reach me at 727-846-0866 or at pohagan@pssnicomp.com to discuss this further. Thank you. Patrick O'Hagan Particle Sizing Systems
Report this Post