Discussion Category: Centrifuge
eppendorf 5810R board diagnostics
56 Replies



Were you able to fix the error 6 on your machine? Please let me know. Also, do you have the service manual?
Many thanks.

Hallo, I have the same problem - error 6. Does anybody have service manual to send me? Thanks!! Reno , My adress E-mail: reno.sime@gmail.com

Hello Reno
LabWrench has a partial User Manual for the Eppendorf 5804 / 5804R / 5810 / 5810R Centrifuges posted on the DOCUMENTS tab of the product page.
Below is a direct link to the User Manual.
http://photos.labwrench.com/equipmentManuals/2402-202.pdf
The error codes are noted on page 41.
Please feel free to contact me if you have any questions.
Respectfully
Ronnie Stanley
Ronnie Stanley
LabWrench Administrator
info@labx.com

Hallo Ronnie,
thank you very much for User manual!
Do you have service manuel too?
Thanks and regards
Reno

HI Reno
No we do not have the Service Manual here. There are many publications on the Eppendorf Website:
Click on the following links
1. Support and ep Services
2. Literature Library
3. Manuals
There you will find a list of products to choose from.
Respectfully
Ronnie
Ronnie Stanley
LabWrench Administrator
info@labx.com

Hello
Sorry for my english .
Usualy the problem ( in the case of error 6 - over load ) it's in a IGBT conected betwen the bridge rectifier and the brake resistor located in the inverter module.
This module it's the module conected to the motor.
sorry my english but today i've repaired another centrifuge with the same problem.
Please don,t hesitate to contact me.
All the best
Augusto hospassis@gmail.com

Hi,
I replaced two biggest electrolytic capacitor on frequency convertor board - it solved problem. (there was error 6)
regards Reno Zagreb

Hi Guys
I have a fault with this centrifuge with lid lock issue. When trying to close the top lid, the motor tries to latch the lid but stops half way. after few tries it does close, but fails to open the lid when button is pressed. I have replaced the lid lock motor and the lid microswitches but fault remains. I have monitored the motor voltage. there motor works without load, but voltage cuts off as soon as load is applied to it.
any suggessions on what might be causing the motor voltage to cut off??
also link to service manual would be helpfull.
thanks
Sarwan