Discussion Category:  Centrifuge

eppendorf 5810R board diagnostics

Error 6 has numerous causes, this unit was working until moved, now it's all error 6. Temp control works fine. Does anybody know if the service manual is worth the $$? I don't want to replace the freq conv bd if I can see a work around. I have several other svc manuals for eppendorf, but the tech support folks are so unhelpful, that I may just recommend replacing unit with another more servicable brand.
avatar placemark

Asked by

AnonymousUser
Report this Post

62 Replies

SimbaReply by
Simba

Have you tried removing the seal?  If you have... yes I would suspect the motor is the issue. 

Report this Post
jchaoReply by
jchao

I have another old model 5810. The sponge inside the lid starts to degrade and spread the black sticky stuffs inside the chamber. My "SMART" co-worker come up with a solution to seal the top hold to prevent these debris. Then the unit went overheat. Since then, It always show overload error when running high speed with load (4000 rpm with buckets). It will be fine without buckets. I pressed UP and DOWN button at the same time, it shows 9:Er 6 2 07. I have replaced Frequency converter twice, and the brake resister once (both show 150 ohms). Nothing works. Is it time to try replace motor?

Report this Post
avatar placemarkReply by
GSPell

For an Error 6, I have had luck just reseating connectors on the main board.

It does not always work, though. The Error 6 I am working on now has a solder run blown completely off the board

Report this Post
avatar placemarkReply by
fidocklab

Hi,

the 5810R in my lab suddenly stopped working with the Error 6. the error shows "9:6 0 1 00" when i press the up and down arrows at the same time.

eppendorf says they do not support repairs or parts for this model anymore.

is there any way to fix it ? or do we need to get a new one but thats really expensive..

thanks,

Sach

Report this Post
avatar placemarkReply by
jayk

Hello!

I have an Eppendorf 5804R that is giving error 24.

The supplier recommended me to change the whole board, but I am actually looking for a more economical solution.

Anyone can help me on that error?

Thank you


Jay
Report this Post
jchaoReply by
jchao

I have an old 5810R. When power it on, it always ask me to press open lid. Once I open and close the lid, it gives error 2 immediately. I've adjust the imbalance sensor to 667 followed by the service manual, the problem still exist, re-check the balance sensor, still with in the spec 670+/-30.

 

Report this Post
avatar placemarkReply by
hxl

Glad to see the question here because I have suffered this once before. Now I know how to make it. Biotech is always complicated for me. Thanks! 


Enjoy the Life!
Report this Post
avatar placemarkReply by
marcot

Awesome! Thanks a lot for that Service Manual. It just helped us to get a centrifuge running again after an "Error 2" after we had to remove a lot of rust and dirt out of an old centrifuge!

Adjusting the imbalance-sensor still is a huge pain! :)

 

best, Marco

Report this Post
avatar placemarkReply by
mforrest

Hi, jjj

Could you send me a link to the old service manual too?

Thanks LF

mforrest at ku dot edu

Report this Post
avatar placemarkReply by
service3683

Dear Alexan. Your error is related to imbalance sensor. Check the brown-yellow-green-white ribbon cable connections at the board. You can take it out and in, clean the pins. If not solved do the following: with the centrifuge lid open push both arrow keys, display will change. press up until you see as first digit "6:".  You should have a value of 670 +/- 100 there. If you move the rotor axis this value should change. If the value displayed is out of range, the imbalance sensor should be readjusted.

please check this and post your results in order to help you further.

Report this Post

Page 4 of 7