Discussion Category:  Preparative HPLC

Calibration

What are the standard used for calibrating the UV detector, fluorescence detector and Refractive index detector. And what column can be used for calibration.

cloudyarun

Asked by

cloudyarun
Report this Post
LabWrenchEditorReply by
LabWrenchEditor

We Have 10 Dionex Hplc And I Do Calibration And Iq Oq And Pq For All Our Instrument

For Uv Detector Use Caffine , Erbium Perchlorate , Cobalt Sulphate
Caffine And Uracil For Resolution.

Thank you to Nazih Asfour for providing this response to LabWrench via LinkedIn


Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com


Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
 
Report this Post
LabWrenchEditorReply by
LabWrenchEditor

EDQM published equipment qualification guidelines for all types equipments used in pharma laboratories. All instruments should pass these criteria and follow recommended conditions. You can reach in the link given below.

Best Regards,
irfan

http://www.edqm.eu/en/quality-management-guidelines-86.html

Thank you to Irfan for providing this response to LabWrench via LinkedIn


Kylie Sauve
LabWrench Site Manager
kylies@labwrench.com


Stay Connected with LabWrench:
Follow us on Twitter
Like us on Facebook
Join the Lab Equipment Solutions Group on LinkedIn
 
Report this Post

Page 1 of 1