air value
the carbo qc show air error then how did we fill the air value
Discussion Category: Chemistry Analyzers
the carbo qc show air error then how did we fill the air value
Asked by
If the CarboQC shows an error for the air value after analysis there is usually a leak on either the inflow or outflow valve. Can you undertake an analysis (measurement) cycle and record the pressure when the outflow valve closes i.e. when liquid stops flowing, when the piston stops for the first time and when the piston stops for the second time and then post the three values for me to look at.