Fatal Error 404
Having a problem with our Molecular Devices SpectraMax Plus 384 with a fatal error 404 or 303 every time we turn on the instrument.
Discussion Category: Software
Having a problem with our Molecular Devices SpectraMax Plus 384 with a fatal error 404 or 303 every time we turn on the instrument.
Reply by303 unable to cal dark current
404 Plate NVRAM air cal data error
So what exactly is the NVRAM air cal data error?
Thanks,
Rich Stahl
Reply by303 is caused by too much stray light and 404 is a cascade error caused by a different error, I have to look in the instrument memory bank to determined what it is.
These errors are not anything that can be fixed by end user. You have to send the unit in for repair and or have a technician come to your place to look at it.
I am the owner and president of A&K Biosource and have many years of experience with Molecular Devices readers I'll be glad to take care of it for you. Visit our website for more info and contact me.
How to deal with the "fatal error 303'?
We have the same problem, "fatal error 303 " was displayed on the screen, when I turn on the devide. How did you fix the device?
Reply byThese instruments and errors are not the kind that can be fixed remotely over the phone or through emails. You need special; tools and software to work on them. I'll be glad to look in to it for you. A&K Biosource, Inc