Discussion Category:  Hematology Analyzers

Pressure/Vacuum error

The pressure unit unable to adjust to the required value (0.039-0.069). The knob after turning did not respond on the screen. (The highest value got was 0.009mmHg) Please what I do to resolve this new machine(F series).

avatar placemark

Asked by

Oluwasegun
Report this Post

2 Replies

avatar placemarkReply by
CA Bryant

This will require a service call.  It sounds like you have a leak somewhere.  It should be an extremely rare occasion when you have to adjust the pressure or vacuum, and only after talking to service first.

Report this Post
avatar placemarkReply by
Van Tien

you need check membrance of pump , and listen sound of pressure .

Report this Post

Page 1 of 1