Discussion Category:  Moisture Analyzers

Not stable-Filtering

balance unstable in environment Customer has two other hr-73 balances work just fine on medium filter setting.This balance won't calibrate due to instability even on high filter setting Is there a way to increase filtering maybe pot on board or resistor or service settings? This balance will calibrate when moved to a more stable environment.Any help would be appreciated. Thanks DT

avatar placemark

Asked by

ksbserv
Report this Post
jburkman11Reply by
jburkman11

Hello,

Environment is everything.  There is also a vibration adapter VIBRA:High setting.  If it is unstable with this setting, the unit should be relocated.

Another option would be to place the HR73 on a marble or granite base which could mitigate some of the vibration effects.

Thanks,

Jennifer


Jennifer Burkman
METTLER TOLEDO
www.mt.com  