Discussion Category:  GC Systems

not ready

How can I change gc status from >not ready< to >ready<
The machine is giving the comment >nrdy< while on the computer is >ready<
when pressing start the sample that should be run has status as already done...

 

avatar placemark

Asked by

Monia
Report this Post
avatar placemarkReply by
Carl M

Depress the "status" key should tell you what heater, valve, detector or peripheral that is not ready.  You may have to adjust an autosampler setting via autosampler key or enter the gc configuration table via gc configure key to set for example the equilibration time.

Report this Post
LabWrenchSiteAdminReply by
LabWrenchSiteAdmin

Hi Mona

It could be a communication problem, you should have the same status on your computer than on your GC, reboot your GC before your computer and then check if comm is ok. Is it just the diode behind the GC oven door wich has a not ready status?

Thanks to Manuel M for providing this response to LabWrench via LinkedIn


Ronnie Stanley
LabWrench Administrator
info@labx.com  
Report this Post
LabWrenchSiteAdminReply by
LabWrenchSiteAdmin

When you reboot try turning on the computer first and open your Method then turn on the GC.

Thanks to Sherry G for providing this response to LabWrench via LinkedIn.


Ronnie Stanley
LabWrench Administrator
info@labx.com  
Report this Post
LabWrenchSiteAdminReply by
LabWrenchSiteAdmin

See wat is the initial temp of ur method on gc through harware....
Some times the harware condition became duffer from wat u enter through ur software ....

Thus this kind of prob occures

Thanks to Preetika S for providing this response to LabWrench via LinkedIn.


Ronnie Stanley
LabWrench Administrator
info@labx.com  
Report this Post
LabWrenchSiteAdminReply by
LabWrenchSiteAdmin

Hi...ok it is bad problem some time you need wait the GC to be ready may be that flow is not stable so you must check it by flowmeter (digital) and compare it with flow in GC also check (split flow) and cylinder pressure 4 bar is suitable

Thanks to Ahed S for providing this response to LabWrench via LinkedIn.


Ronnie Stanley
LabWrench Administrator
info@labx.com  