Humalyte plus 3 problem

the results of my analyzer humalyt e plus 3 are not stable

i have changed the reference electrode and the ISE values become in the range but still very low and the results are very far than expected

please help me

avatar placemark

Asked by

maham
Report this Post
avatar placemarkReply by
Rosinaldo

Ola, vc precisa fazer uma repetitividade ou reprodutibilidade , passe 5 Vezes uma mesma Amostra e anota os Resultados Na, K, Cl e : Não pode variar Mas Que 5% para cada Parâmetro, se estiver Dentro Dessa variação vc Só Precisa Fazer uma Calibração de Fatores use um Soro Controle de bioquimica da Marca Que vc Confia, eu uso da Labtest e uso Como referencia Valores os valores Ion Seletivo AVL da roche pois o humalyte plus 3 e o 5 São EQUIPAMENTOs relativamento Novos e os  Fabricantes de soro Controle ainda Não anexaram às Tabelas de Referência para este EQUIPAMENTO. Mais lhe adianto que é  a mesma do  AVL da Roche. somente a Human Invitro fibricante do equipamento tem  Soro Controle com Valores de Referência expecíficos pra este equipamento. Caso a reprodutibilidade esteja fora dos 5% Precisa de Manutenção corretiva ou preventiva UO para verificar os Problemas. qualquer duvida Contato Entre em contato. E-mail. rpbarbosa2006@gmail.com  sou técnico científo, faço manutenção em equipamentos laboratoriais e hospitalares. trabalho no distribuidor da Human Invitro do Pará e já fiz treinamento na human Invitro de itabira-MG. abraços qualquer coisa entre em contato.  

TRANSLATION:

Message:

Hello, u need to make a repeatability or reproducibility, spend 5 times the same sample and note the results Na, K, and Cl: But what can not vary 5% for each parameter, if vc Within this variation Just Need to Make a Calibration Factors use a Control Serum biochemical Brand That you trust, I use the Labtest How to use reference values ??and the values ??of Ion Selective AVL roche because humalyte plus 3 and 5 are equipment relativamento New and Manufacturers Control Serum not yet attached to Reference Tables for this EQUIPMENT. More I advance you that is the same as the Roche AVL. Only Human fibricante Invitro equipment has Control Serum with Reference Values ??expecíficos to this equipment. If reproducibility is 5% off the Need Maintenance corrective or preventive OU to check problems. Contact any questions please contact. E-mail. rpbarbosa2006@gmail.com 'm científo technician, do maintenance on laboratory and hospital equipment. work in Invitro Human distributor of Para and I've done the training on human itabira Invitro-MG. Hugs anything contact.

Report this Post
avatar placemarkReply by
labotronics.mary
Unstable and unusually low results on the Humalyte Plus 3 may be caused by clogged ISE tubing, air bubbles, contaminated electrodes, incorrect calibration, or problems with the reagents. Since replacing the reference electrode has not resolved the issue, check the electrode connections, flow path, and reagent expiry, then perform cleaning and calibration according to the manufacturer’s manual. Run quality-control samples to verify accuracy, and contact a qualified service engineer if the results remain unstable. Avoid reporting patient results until the analyzer passes quality-control checks.
Report this Post