Discussion Category:  Titrators

PM and Calibration

Hello,

 

Can you please help me out to solve a mistery related to the required frequency in terms of PM and/or calibration for a T50 unit? My question is related to the equipment itself and not to the electrod. Is there a required frequency from manufacturer?

 

Thanks a lot,

Angela

avatar placemark

Asked by

Marie Curie
Report this Post
jburkman11Reply by
jburkman11

Hello,

There is no required frequency from the manufacturer, this is strictly according to the customer's SOP.  But, many customers follow these guidelines: PM's are generally done once or twice a year, calibration of pH electrodes would be done daily if a pH endpoint calibration is being done, titrants are "calibrated" (titer done) every two weeks or when the titrant bottle is replaced.

Thanks,

Jennifer


Jennifer Burkman
METTLER TOLEDO
www.mt.com  
Report this Post
avatar placemarkReply by
nfabius

Could I know which items are revised during the PM, please?

Report this Post
jburkman11Reply by
jburkman11

Specfic details are not available. PM service does include;

General technical instrument inspection, cleaning and lubrication of moving parts, alignment correctness of sample positions, operational verification of perpherials, function verification of interfaces.

 


Jennifer Burkman
METTLER TOLEDO
www.mt.com  
Report this Post