CP-3800 GC SampleList will not run automatically

I have a CP-3800 GC with CP-8410 AutoInjector run by Star Chromatography Workstation 6.41. I tried to run SampleList but will not work automatically. I have to press "START" for each sample individually. Otherwise, the Autoinjector will roll over the SampleList but will not inject. For each sample (analysis) it will wait for sometime to be STARTed. If not started, it will move on to the next sample...Any suggestions?

avatar placemark

Asked by

Kamir
Report this Post
avatar placemarkReply by
ArielC

Dear ,

Please you go to system - configuration and see if you have the GC´s icon inside of "autostart" square.

 

Regards

 

Ariel


Chromatography Specialist. Varian and YL Instrument.

http://www.labsolutions.com.ar

   
Report this Post
avatar placemarkReply by
Kamir

Hello,

Thank you for the tip. It worked perfectly after the configuration had been changed.

Best regards.