Discussion Category:  MS Detectors

CI/EI SWITCH WANTED?

THE PRESSURE IS TOO HIGH AND THE MESSAGE WAS CODE 520 HV SUPPLY FAIL  AND NO CURRENT EMISSION . WHAT DO I NEED TO DO?

avatar placemark

Asked by

Etrevino
Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie

This can be the symptom of a minor problem such as insufficient vacuum or a much larger problem.  Do you have adequate vacuum read from a high vacuum gauge (not the pirani gauge on the low vacuum side)?

 

Dan


Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post
avatar placemarkReply by
labotronics.mary
The high pressure may be preventing the ion source from operating correctly and triggering the “Code 520 HV Supply Fail” error with no emission current. First, check the foreline pump, vacuum connections, and possible leaks according to your instrument manual. Once the vacuum reaches the specified operating level, check the ion source and high-voltage supply. Avoid adjusting high-voltage components yourself; if the error persists, contact a qualified service technician and provide the instrument model.
Report this Post

Page 1 of 1