Discussion Category:  MS Detectors

CI/EI SWITCH WANTED?

THE PRESSURE IS TOO HIGH AND THE MESSAGE WAS CODE 520 HV SUPPLY FAIL  AND NO CURRENT EMISSION . WHAT DO I NEED TO DO?

avatar placemark

Asked by

Etrevino
Report this Post
Dan Ste.MarieReply by
Dan Ste.Marie

This can be the symptom of a minor problem such as insufficient vacuum or a much larger problem.  Do you have adequate vacuum read from a high vacuum gauge (not the pirani gauge on the low vacuum side)?

 

Dan


Dan Ste.Marie

RJ2 Instrument Services

Erie, PA

www.rj2.biz

   
Report this Post