Discussion Category: Centrifuge
error99
i wanna reset my allegra x-22, how can i do that?

I guess that my question for you is why you believe that you need to reset your centrifuges control board? What is the issue that you are trying to resolve?
Don

OK, I don't have a manual handy to me at the moment so what does a Code 99 tell you in the Operator's Manual?
Don

Here is error 98 & 99, both have to do with rotor ID.
98
rotor ID
invalid rotor ID
rotor ID sensor to rotor gap, rotor missing magnets, rotor ID sensor and cable
99
rotor ID
invalid rotor ID
wrong rotor ID keyed in by user, same as error #98

Craig, Well, if your error code indicates a problem with the rotor ID then that's exactly what the problem is and resetting the centrifuge control board will not fix the problem. I have to ask the obvious, is the rotor code that's programmed into the centrifuge the same as what the rotor itself is? If not, then set it properly. If it is then you will probably need to replace the rotor ID sensor. It's a very simple system so there's not much else to say about it.
Don